Signal reception

Hi. I’m having some troubles getting the signal from the MRI scanner.
When I use the lumina 3G controller in ASCII mode, it works fine, but in eprime mode the signal delays and it is received only each 12 dynamic scans. It happens only when the MRI is running, but not when testing the response pads without the MRI running. I’d like to know if it could be something from the controller or if you have any suggestion. Thanks

Please download the Xidon 2 software from

https://cedrus.com/support/xid/xidon.htm

When you run it and it detects your Lumina controller, it will display the firmware version. Which one do you have?

The firmware version I have is v2.0.2

Please see this page, Updating the Controller’s Firmware. This issue was fixed in firmware in version 2.0.6.