Signal reception, Posted December 7th, 2018 #1
Hi. I'm having some troubles getting the signal from the MRI scanner.
When I use the lumina 3G controller in ASCII mode, it works fine, but in eprime mode the signal delays and it is received only each 12 dynamic scans. It happens only when the MRI is running, but not when testing the response pads without the MRI running. I'd like to know if it could be something from the controller or if you have any suggestion. Thanks
Reply With Quote
Hisham
Founder
 
Join Date: Aug 2006
Location: San Pedro, CA
Hisham is on a distinguished road Posts: 1,216
Posted December 10th, 2018 #2
Please download the Xidon 2 software from
https://cedrus.com/support/xid/xidon.htm
When you run it and it detects your Lumina controller, it will display the firmware version. Which one do you have?
Reply With Quote
bldth
Junior Member
 
Join Date: Jul 2018
bldth is on a distinguished road Posts: 4
Posted December 13th, 2018 #3
The firmware version I have is v2.0.2
Reply With Quote
Hisham
Founder
 
Join Date: Aug 2006
Location: San Pedro, CA
Hisham is on a distinguished road Posts: 1,216
Posted December 13th, 2018 #4
Please see this page, Updating the Controllerís Firmware. This issue was fixed in firmware in version 2.0.6.
Reply With Quote